The Accuracy of the FreeStyle Navigator Continuous Glucose Monitoring System in Children With Type 1 Diabetes
Darrell M. Wilson, MD, Roy W. Beck, MD, PHD, William V. Tamborlane, MD, Mariya J. Dontchev, MPH, Craig Kollman, PHD, Peter Chase, MD, Larry A. Fox, MD, Katrina J. Ruedy, MSPH, Eva Tsalikian, MD, Stuart A. Weinzimer, MD and the DirecNet Study Group
From the DirecNet Coordinating Center, Jaeb Center for Health Research, Tampa, Florida
Address correspondence and reprint requests to Darrell M. Wilson, MD, DirecNet Coordinating Center, Jaeb Center for Health Research, 15310 Amberly Dr., Suite 350, Tampa, FL 33647. E-mail: direcnet@jaeb.org (direcnet@jaeb.org)
OBJECTIVE—To evaluate the accuracy and precision of the FreeStyle Navigator continuous glucose monitoring system in children with type 1 diabetes.
RESEARCH DESIGN AND METHODS—In 30 children with type 1 diabetes (mean age 11.2 ± 4.1 years), the Navigator glucose values were compared with reference serum glucose values of blood samples obtained in an inpatient clinical research center and measured in a central laboratory using a hexokinase enzymatic method and in an outpatient setting with a FreeStyle meter. Median absolute difference (AD) and median relative absolute difference (RAD) were computed for sensor-reference and sensor-sensor pairs.
RESULTS—The median AD and RAD were 17 mg/dl and 12%, respectively, for 1,811 inpatient sensor-reference pairs and 20 mg/dl and 14%, respectively, for 8,639 outpatient pairs. The median RAD between two simultaneous Navigator measurements (n = 1,971) was 13%. Ninety-one percent of sensors in the inpatient setting and 81% of sensors in the outpatient setting had a median RAD http://care.diabetesjournals.org/math/le.gif20%.
CONCLUSIONS—The Navigator’s accuracy does not yet approach the accuracy of current-generation home glucose meters, but it is sufficient to believe that the device has the potential to be an important adjunct to treatment of youth with type 1 diabetes.
Abbreviations: AD, absolute difference • CGMS, continuous glucose monitoring system • CRC, clinical research center • ISO, International Organisation for Standardization • RAD, relative absolute difference
Emerging Treatments and Technologies
Original Article
The Accuracy of the FreeStyle Navigator Continuous Glucose Monitoring System in Children With Type 1 Diabetes
Darrell M. Wilson, MD, Roy W. Beck, MD, PHD, William V. Tamborlane, MD, Mariya J. Dontchev, MPH, Craig Kollman, PHD, Peter Chase, MD, Larry A. Fox, MD, Katrina J. Ruedy, MSPH, Eva Tsalikian, MD, Stuart A. Weinzimer, MD and the DirecNet Study Group
From the DirecNet Coordinating Center, Jaeb Center for Health Research, Tampa, Florida
Address correspondence and reprint requests to Darrell M. Wilson, MD, DirecNet Coordinating Center, Jaeb Center for Health Research, 15310 Amberly Dr., Suite 350, Tampa, FL 33647. E-mail: direcnet@jaeb.org (direcnet@jaeb.org)
OBJECTIVE—To evaluate the accuracy and precision of the FreeStyle Navigator continuous glucose monitoring system in children with type 1 diabetes.
RESEARCH DESIGN AND METHODS—In 30 children with type 1 diabetes (mean age 11.2 ± 4.1 years), the Navigator glucose values were compared with reference serum glucose values of blood samples obtained in an inpatient clinical research center and measured in a central laboratory using a hexokinase enzymatic method and in an outpatient setting with a FreeStyle meter. Median absolute difference (AD) and median relative absolute difference (RAD) were computed for sensor-reference and sensor-sensor pairs.
RESULTS—The median AD and RAD were 17 mg/dl and 12%, respectively, for 1,811 inpatient sensor-reference pairs and 20 mg/dl and 14%, respectively, for 8,639 outpatient pairs. The median RAD between two simultaneous Navigator measurements (n = 1,971) was 13%. Ninety-one percent of sensors in the inpatient setting and 81% of sensors in the outpatient setting had a median RAD http://care.diabetesjournals.org/math/le.gif20%.
CONCLUSIONS—The Navigator’s accuracy does not yet approach the accuracy of current-generation home glucose meters, but it is sufficient to believe that the device has the potential to be an important adjunct to treatment of youth with type 1 diabetes.
Abbreviations: AD, absolute difference • CGMS, continuous glucose monitoring system • CRC, clinical research center • ISO, International Organisation for Standardization • RAD, relative absolute difference